U/T rep rate

We test nickel alloy billet material using basic pulse echo techniques. our method spec calls for a minimum pulse repetition rate of 500PPS. We often experience "wraparound" at this rate and have to lower the rep rate to eliminate the unwanted signals. My question is, why is it so important to maintain that rep rate? What harm could dropping it to 450 - 400 PPS do? Max surface scanning speed is 10 lin. inches per/sec. Transducer is 10 or 5MHz. Beamwidth as measured at .150" metal travel on a #1 FBH cal block is @ .040".


 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 183

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
183
06:46 Dec-17-2002
Re: U/T rep rate
reply:
Hello Terry,

If your scanning speed v is 10"/s and your beam width b is only 0.040", i.e. per shot you cover only 0,040", you need a pulse repetition rate for complete coverage of
frep = v/b.
For your example that means frep = (10"/0.04"s)=250Hz at least.
If you want a factor n for redundancy, i.e. n pulses per beam diameter b, the equation goes to
frep = n v/b.
In your case n=2 means frep = 2 10"/.040"s) = 500 Hz.

You are correct, that this repetition frequency may be too high for low attenuating materials as nickel based alloys are.
The only solution then is to look for the maximum repetition rate fmax you can work with (static), and for scanning to lower the scanning speed until you get full coverage including a safety factor n:
vmax = (b fmax/n)
If you get fmax = 50 Hz, with n = 2 you will get the maximum allowable scanning speed vmax = (0.04" 50/2s)= = 1"/s only.
May be, for lower repetition rates you need a digital ultrasonic instrument, because they have separate screen refreshing rates independent from pulse repetion frequency, whereas with analog instruments the screen refreshing rate is the same as the pulse repetition frequency and for low rates you will not see a constant screen display.

Kind regards
Udo Schlengermann
Agfa NDT GmbH, Huerth, Germany
+++

: We test nickel alloy billet material using basic pulse echo techniques. our method spec calls for a minimum pulse repetition rate of 500PPS. We often experience "wraparound" at this rate and have to lower the rep rate to eliminate the unwanted signals. My question is, why is it so important to maintain that rep rate? What harm could dropping it to 450 - 400 PPS do? Max surface scanning speed is 10 lin. inches per/sec. Transducer is 10 or 5MHz. Beamwidth as measured at .150" metal travel on a #1 FBH cal block is @ .040".
.



 
 Reply 
 

Product Spotlight

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Echomac® PA Tube Weld Tester

The new robotic Echomac® PA TW is an operator friendly test system for detecting weld zone defects
...
in tube and pipe caused by scarfing or welding processes. Electronic scanning ensures that the entire weld zone is covered without the need for mechanical movement or operator adjustment.
>

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

PAUT Probes

Typical Phased Array probes have frequencies between 1MHz and 20MHz and the number of wafers is 10
...
to 128. M2 Electronics offers customers conventionally ultrasound probes and the ability to provide high-precision Phased Array Ultrasound Probes of up to 256 wafers. We can also customize the probe for our customers to meet the specific application requirements of the user.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window