MPT bench current meters

Can anyone explain the industry situation as regards the problem below? (NDT is not my field, but I am editing/writing some vocational-level educational materials on MPT.)

THE PROBLEM:
Lets say you want do follow an MPT method - eg for testing some component of standard dimensions - developed elsewhere, which specifies a magnetising current. You want to generate the same current with your unit.
Now the MPT test bench at out college (a "lectro-flux BWD 1565") has a simple moving coil meter, and an AC-DC selector. The machine is old, and I dont have access to the manuals.
There is no indication of whether the meter is indicating RMS or peak, and no indication of whether "DC" means half- or full-wave rectified. In addition, even if I knew the meter was indicating RMS for an AC current, can I trust the manufacturer to have included circuit to account for the change in form factor, so that the current reading for a "DC" current is also true RMS? Or is the reading on DC really meaningless - for comparison purposes only?
My questions are:
1)Can one trust the indicated values from moving coil meters on MPT benches to be true RMS ?
2) Is is a typical situation to not know the "true" value of the current, except to be able to compare a current with others generated on the same machine ?
3) Is it rare situation to want to carry out a procedure developed on a different machine? Or must it be re-validated on every machine?
4) Does anyone have any horror stories stemming from blindly believing a quoted current in a procedure, or blindly believing the current reading of an MPT bench ?





    
 
 
Neil Burleigh
Sales
Krautkramer Australia Pty Ltd, Australia, Joined Dec 2002, 152

Neil Burleigh

Sales
Krautkramer Australia Pty Ltd,
Australia,
Joined Dec 2002
152
02:58 Feb-05-2003
Re: MPT bench current meters
: Can anyone explain the industry situation as regards the problem below? (NDT is not my field, but I am editing/writing some vocational-level educational materials on MPT.)
.
: THE PROBLEM:
: Lets say you want do follow an MPT method - eg for testing some component of standard dimensions - developed elsewhere, which specifies a magnetising current. You want to generate the same current with your unit.
: Now the MPT test bench at out college (a "lectro-flux BWD 1565") has a simple moving coil meter, and an AC-DC selector. The machine is old, and I dont have access to the manuals.
: There is no indication of whether the meter is indicating RMS or peak, and no indication of whether "DC" means half- or full-wave rectified. In addition, even if I knew the meter was indicating RMS for an AC current, can I trust the manufacturer to have included circuit to account for the change in form factor, so that the current reading for a "DC" current is also true RMS? Or is the reading on DC really meaningless - for comparison purposes only?
: My questions are:
: 1)Can one trust the indicated values from moving coil meters on MPT benches to be true RMS ?
: 2) Is is a typical situation to not know the "true" value of the current, except to be able to compare a current with others generated on the same machine ?
: 3) Is it rare situation to want to carry out a procedure developed on a different machine? Or must it be re-validated on every machine?
: 4) Does anyone have any horror stories stemming from blindly believing a quoted current in a procedure, or blindly believing the current reading of an MPT bench ?
.
.Paul,
Give Dave Mullet a call at Lectromax Australia 03 94311041 admin@lectromax.com.au. I think they made the mag bench you are talking about.
Also I always recommend that mag flux indicators be used on the test piece because you cannot guarantee the output of the test bench is transferred completely into the test piece.

Regards
Neil Burleigh


    
 
 

Product Spotlight

Digital Radiography for Concrete Inspection

In the construction industry (buildings, bridges) there is a need for fast non- destructive means,
...
most commonly to locate targets (Pipes, Re-bars, Conduit, Fittings etc.) within concrete structures prior to drilling, cutting or coring. Measuring reinforcements is also performed for static calculations and to detect corrosion problems and damages. Cracks and voids can also be visualized. The examinations are not only restricted to concrete constructions; they can also be applied to reinforced masonry made of natural stones or bricks.
>

Aerospace Systems - Automated Ultrasonic Inspection

USL are specialists in the design and manufacture of turnkey ultrasonic inspection systems for aer
...
ospace applications. From monolithic composites to complex honeycomb structures. This video shows just a few examples of what is possible, find out more at: www.ultrasonic-sciences.co.uk
>

High Resolution Industrial CT System for Small/Medium Size Parts Inspection

The YXLON FF35 CT computed tomography system is designed to achieve extremely precise inspection res
...
ults for a wide range of applications. Available in a single or dual tube configuration, it is perfect for very small to medium size parts inspection in the automotive, electronics, aviation, and material science industries.
>

Research Center IDEKO develops an ultrasonic train wheel inspection system for CAF

The Basque Research centre IK4-IDEKO has developed a state-of-the-art ultrasonic inspection system f
...
or the train wheel. This system secures a sound condition of train wheels and is thus a significant contribution to rail transport safety. The device was delivered to CAF recently and its use allows the manufacturer to become an approved supplier of rolling stock in Italy, as it meets the demanding homologation standards of this country. \\\\r\\\\n
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window
s