MM Damani
MM Damani
00:00 Feb-12-2003
Hidden Crack in Graphite Electrodes

USE OF ACOUSTIC RESONANCE FOR DETECTING HIDDEN CRACKS IN GRAPHITE ELECTRODES”


Q.1 Whether Acoustic Resonance method uses mechanical waves having sinusoidal oscillations?

Q.2 Whether the Instrument is capable to detect cracks, extrusion defects, cavities, porosity etc. in Carbon / Graphite Electrodes?

Q.3 If yes what the level of Precision and repeatability of Test?

Q.4 Whether it is possible to check carbon / Graphite Cylindrical Rods of diameter ranging between 250 mm to 700 mm and length between 1800 mm to 3000 mm?

Q.5 Is it possible to carryout Acoustic resonance test on Pitch Impregnated Carbon Rods for ascertaining pitch pickup level after impregnation?



    
 
 
Godfrey Hands
Engineering,
PRI Nadcap, United Kingdom, Joined Nov 1998, 289

Godfrey Hands

Engineering,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
289
01:49 Feb-14-2003
Re: Hidden Crack in Graphite Electrodes
Hi There,

My company uses the Quasar Resonant Inspection equipment for inspecting components. This uses a sinusoidal wave to excite the parts mechanically.

We are normally able to detect Cracks, Extrusion defects, Cavities, Significant amounts of Porosity and many other conditions in Metal and Ceramic components, as well as Ferrite. To my knowledge, we have never tested Graphite electrodes, but as the technology is quite new, it will only be a matter of time before it is tried on Graphite.

We normally measure resonances with a precision of "a few Hertz" for optimising testing speed, but the system has a resolution capability of 1 Hz. Our measurement errors (repeatability) are typically 0.01% or smaller at higher frequencies, and perhaps 0.05% at lower frequencies.

We normally test components in the weight range 20 Grams to 20 Kilograms, but are currently experimenting with components around 50 to 70 Kg. I see no reason why the technology shouldn't test your electrodes.

I think that it is unlikely that we will be able to determine the pitch pickup level of pitch impregnated electrodes, but cannot rule this possibility out.

Please provide your location, and I can arrange that your nearest Quasar Application Engineer or Sales Person contacts you to make some trials on some of your electrodes.

Regards,

Godfrey Hands

-------------------------------------
: USE OF ACOUSTIC RESONANCE FOR DETECTING HIDDEN CRACKS IN GRAPHITE ELECTRODES”
.
:
: Q.1 Whether Acoustic Resonance method uses mechanical waves having sinusoidal oscillations?
.
: Q.2 Whether the Instrument is capable to detect cracks, extrusion defects, cavities, porosity etc. in Carbon / Graphite Electrodes?
.
: Q.3 If yes what the level of Precision and repeatability of Test?
.
: Q.4 Whether it is possible to check carbon / Graphite Cylindrical Rods of diameter ranging between 250 mm to 700 mm and length between 1800 mm to 3000 mm?
.
: Q.5 Is it possible to carryout Acoustic resonance test on Pitch Impregnated Carbon Rods for ascertaining pitch pickup level after impregnation?
.



    
 
 
Lalit Chandra
Lalit Chandra
08:03 Mar-27-2015
Re: Hidden Crack in Graphite Electrodes
In Reply to Godfrey Hands at 01:49 Feb-14-2003 .

Dear Sir,

I am working in a graphite manufacturing plant. very much interested to know the results of any test/experiments with graphite cylindrical rods.

if your company did any experiment to detect cracks inside these electrode rods please let us know.

lalit

    
 
 
Godfrey Hands
Engineering,
PRI Nadcap, United Kingdom, Joined Nov 1998, 289

Godfrey Hands

Engineering,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
289
12:29 Mar-27-2015
Re: Hidden Crack in Graphite Electrodes
In Reply to Lalit Chandra at 08:03 Mar-27-2015 .

Hi,
I no longer work directly with this company, but I can get more details sent to you.
Please provide me with contact details to my email address.
Please use the email option in NDT Net.

Regards,

Godfrey Hands

    
 
 
Godfrey hands
Engineering,
PRI Nadcap, United Kingdom, Joined Nov 1998, 289

Godfrey hands

Engineering,
PRI Nadcap,
United Kingdom,
Joined Nov 1998
289
16:31 Jan-19-2017
Re: Hidden Crack in Graphite Electrodes
In Reply to Godfrey Hands at 12:29 Mar-27-2015 .

Hi,
I have recently had another enquiry about testing Graphite Electrodes with this technology.

Back in 2003 when this idea was first considered, this technology was very new, and nobody knew exactly what potential there was for inspecting Graphite electrodes, although it had been considered a possibility.
14 years later, and we now know a lot more about what can be done and what is doubtful, and inspection of large Graphite Electrodes now appears to fall into the "Doubtful" category, although some limited success has been reported for inspecting Graphite components of up to about 2Kg for major defects.

The technology is however still an excellent tool for mass-production of smaller metallic or fired ceramic components.

Godfrey

    
 
 
Krishna Bhuta
Director,
Own, India, Joined Jun 2014, 6

Krishna Bhuta

Director,
Own,
India,
Joined Jun 2014
6
08:34 Jan-23-2017
Re: Hidden Crack in Graphite Electrodes
In Reply to Godfrey hands at 16:31 Jan-19-2017 .

Hi, I had worked on this project some tiem back, its possible to carry out UT inspection fro mass production too if require.

    
 
 
Simon Amallraja
Consultant, Advanced NDT Methods
Anu Consultants, India, Joined Aug 2013, 38

Simon Amallraja

Consultant, Advanced NDT Methods
Anu Consultants,
India,
Joined Aug 2013
38
15:06 Jan-23-2017
Re: Hidden Crack in Graphite Electrodes
In Reply to MM Damani at 00:00 Feb-12-2003 (Opening).

You can use 410D model from Sonatest UK. Since you wanted to check Porosity etc., you can use Air Coupled System for this application as it can give a C Scan image and porosity, other mentioned defects.

If you are in India, you can contact me for further details. www.anuconsultants.net

    
 
 
Dustin
Dustin
12:42 Aug-11-2017
Re: Hidden Crack in Graphite Electrodes
In Reply to MM Damani at 00:00 Feb-12-2003 (Opening).

I had purchased Graphite Electrode RP 350 mm * 1800 MM with 4 TPI nipple plus 3% wherein the there is a serious quality problem. During operation of the electrodes in electric arc furnace, several times slipage happened. Can you kindly know me the technical deficiency in graphite electrodes due to which such slipage is happening.

    
 
 

Product Spotlight

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

TOPAZ® 64: Fully Integrated Portable 64 Channel PAUT Instrument with FMC & TFM

Introducing TOPAZ64, the industry’s most intelligent fully integrated, portable 64-channel phased
...
array ultrasound (UT) instrument. TOPAZ64 combines code-compliant phased array UT with the industry’s most advanced full matrix capture (FMC) and total focusing method (TFM) capabilities. Featuring the highest acquisition frequency in its class, high resolution FMC and a 12” multi-touchscreen, users can easily visualize even the smallest flaws. TOPAZ64 can generate a bipolar pulse that provides more acoustic energy versus previous models for punching through thick components. The result is a portable tool that delivers increased inspection coverage, more accurate signals, and the ability to handle all UT inspections in one package. TOPAZ64 is ideal for challenging applications in transportation, oil and gas, manufacturing and power generation.
>

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window