calibration holes for leak testing

Good Morning all,

Wehave been performing quantitative studies of leak testing using a useful little setup that terminates in an end cap having a hole. We created a number of holes that range from the too-gross (0.03" diameter drilled) to the moderate (0.004" laser drilled) and have been able to document systematic changes in both bubble and acoustic detection, using orifice size, pressure differential, etc. as variables.

The big question: all these holes seem too big compared to the leaks we are asked to chase in real industrial situations. Does anyone have info/ideas on getting small(er) holes for use in our test system? Our next thought is to isolate and capture real leaks into a larger contraption, but that's looking a bit clumsy.

Thanks for your time.


    
 
 
Phil Herman Jr.
Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards, USA, Joined Oct 1999, 79

Phil Herman Jr.

Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards,
USA,
Joined Oct 1999
79
03:28 Jun-20-2003
Re: calibration holes for leak testing
Hello Dave,

PH Tool Reference Standards has machined very small calibration holes for leak testing for tubing mills. Most recent was a series of holes in .028" wall stainless tube. We EDM'd one through-hole per standard, with diameters of .0017, .0039, and .0079". The feedback I recall on the smallest hole was that it was TOO small. Contact me directly for more details.

Best Regards,
Phil Herman Jr.
President/CEO
PH Tool Reference Standards


: Good Morning all,
.
: We have been performing quantitative studies of leak testing using a useful little setup that terminates in an end cap having a hole. We created a number of holes that range from the too-gross (0.03" diameter drilled) to the moderate (0.004" laser drilled) and have been able to document systematic changes in both bubble and acoustic detection, using orifice size, pressure differential, etc. as variables.
.
: The big question: all these holes seem too big compared to the leaks we are asked to chase in real industrial situations. Does anyone have info/ideas on getting small(er) holes for use in our test system? Our next thought is to isolate and capture real leaks into a larger contraption, but that's looking a bit clumsy.
.
: Thanks for your time.
.



    
 
 
James A. Basford
James A. Basford
04:28 Jun-24-2003
Re: calibration holes for leak testing
: Good Morning all,
.
: Wehave been performing quantitative studies of leak testing using a useful little setup that terminates in an end cap having a hole. We created a number of holes that range from the too-gross (0.03" diameter drilled) to the moderate (0.004" laser drilled) and have been able to document systematic changes in both bubble and acoustic detection, using orifice size, pressure differential, etc. as variables.
.
: The big question: all these holes seem too big compared to the leaks we are asked to chase in real industrial situations. Does anyone have info/ideas on getting small(er) holes for use in our test system? Our next thought is to isolate and capture real leaks into a larger contraption, but that's looking a bit clumsy.
.
: Thanks for your time.
.
For many years we have been using 1/16" OD x 0.004" ID tubing welded into demountable flanges as flow restriction elements for calibrated leaks. The smaller gas flow rates needed for our calibrated leaks were obtained by crimping the end of the capillary tubing. Very constant flow rates from 1E-7 to 1E-2 atm cc/sec with gas source pressures from 1 to 20 atm are readily achievable. These are commercially available. I have also
made reliable orifices by drilling #80 holes (0.013" dia) in blind nickel gaskets and swaging them using a punch to get flow rates in the 1E-9 atm cc/sec range. For more information on calibrated leaks and leak testing, visit the web site of the American Vacuum Society, avs.org. Their Recommended Practices documents are be quite useful.


    
 
 
richard collins
richard collins
07:31 Jul-13-2005
Re: calibration holes for leak testing
: : Good Morning all,
Here at H.T.Products ltd we are leak detection specialists and also manufacturers of calibrated reference leaks of all types for most applications.We have a patent on a metal vacuum sintered leak which covers leak rates from typically 10e1 to 10e9 mbarlit/sec. They have the advantage of fast response, can tolerate high pressure differentials and high temperature are available for most gases with calibration traceability to national standards and re-calibration service available.We welcome your enquires.
: .
: : Wehave been performing quantitative studies of leak testing using a useful little setup that terminates in an end cap having a hole. We created a number of holes that range from the too-gross (0.03" diameter drilled) to the moderate (0.004" laser drilled) and have been able to document systematic changes in both bubble and acoustic detection, using orifice size, pressure differential, etc. as variables.
: .
: : The big question: all these holes seem too big compared to the leaks we are asked to chase in real industrial situations. Does anyone have info/ideas on getting small(er) holes for use in our test system? Our next thought is to isolate and capture real leaks into a larger contraption, but that's looking a bit clumsy.
: .
: : Thanks for your time.
: .
: For many years we have been using 1/16" OD x 0.004" ID tubing welded into demountable flanges as flow restriction elements for calibrated leaks. The smaller gas flow rates needed for our calibrated leaks were obtained by crimping the end of the capillary tubing. Very constant flow rates from 1E-7 to 1E-2 atm cc/sec with gas source pressures from 1 to 20 atm are readily achievable. These are commercially available. I have also
: made reliable orifices by drilling #80 holes (0.013" dia) in blind nickel gaskets and swaging them using a punch to get flow rates in the 1E-9 atm cc/sec range. For more information on calibrated leaks and leak testing, visit the web site of the American Vacuum Society, avs.org. Their Recommended Practices documents are be quite useful.
.



    
 
 
richard collins
richard collins
07:34 Jul-13-2005
Re: calibration holes for leak testing
: : Good Morning all,
Here at H.T.Products ltd we are leak detection specialists and also manufacturers of calibrated reference leaks of all types for most applications.We have a patent on a metal vacuum sintered leak which covers leak rates from typically 10e1 to 10e9 mbarlit/sec. They have the advantage of fast response, can tolerate high pressure differentials and high temperature are available for most gases with calibration traceability to national standards and re-calibration service available.We welcome your enquires.
: .



    
 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window