Technical Discussions
Akshay Singh
Akshay Singh
07:24 Aug-04-2003
PZT film sensing

Hello:

I am using piezo films for sensing cracks in a substrate. I am having problems with dealing with the 60 Hz supply. It is appearing along with its harmonics in the entire frequency spectrum that I am looking at. I shielded the cables with Al foil, covered the film with Al foil but I am still picking up the 60 Hz. I would highly appreciate any suggestions. Regards.. Akshay


 
 Reply 
 
Joerg Schulze-Clewing
Joerg Schulze-Clewing
00:17 Aug-05-2003
Re: PZT film sensing
Dear Akshay,

You'd have to provide some more info about your setup,
such as what frequency range you are using or maybe
which equipment.

Shielding doesn't have much effect at low frequencies
and it is usually better to use coupling transformers
that pass the desired frequency range and isolate
conductive paths. If harmonics are bothering you up
into the frequency range you are using then check for
noise "polluters" on the mains supply. I found one in
our lab yesterday and it was a plain old light dimmer
that had gone bad. It was showing up at several MHz
and after removing it that noise completely vanished.

All the best, Joerg.

----------- Start Original Message -----------
: Hello:
: I am using piezo films for sensing cracks in a substrate. I am having problems with dealing with the 60 Hz supply. It is appearing along with its harmonics in the entire frequency spectrum that I am looking at. I shielded the cables with Al foil, covered the film with Al foil but I am still picking up the 60 Hz. I would highly appreciate any suggestions. Regards.. Akshay
------------ End Original Message ------------





 
 Reply 
 
Akshay Singh
Akshay Singh
07:41 Aug-05-2003
Re: PZT film sensing
Dear Joerg:

Thanks much for the response. You know what? It turned out the surge I was using had a problem with one of the outlet...:-). As soon as I plugged it in a different outlet...BAM...the 60 Hz died. As they say, more than often...the solutions are simple...Thanks again....Akshay

----------- Start Original Message -----------
: Dear Akshay,
: You'd have to provide some more info about your setup,
: such as what frequency range you are using or maybe
: which equipment.
: Shielding doesn't have much effect at low frequencies
: and it is usually better to use coupling transformers
: that pass the desired frequency range and isolate
: conductive paths. If harmonics are bothering you up
: into the frequency range you are using then check for
: noise "polluters" on the mains supply. I found one in
: our lab yesterday and it was a plain old light dimmer
: that had gone bad. It was showing up at several MHz
: and after removing it that noise completely vanished.
: All the best, Joerg.
: : Hello:
: : I am using piezo films for sensing cracks in a substrate. I am having problems with dealing with the 60 Hz supply. It is appearing along with its harmonics in the entire frequency spectrum that I am looking at. I shielded the cables with Al foil, covered the film with Al foil but I am still picking up the 60 Hz. I would highly appreciate any suggestions. Regards.. Akshay
------------ End Original Message ------------




 
 Reply 
 
Joerg Schulze-Clewing
Joerg Schulze-Clewing
01:22 Aug-06-2003
Re: PZT film sensing
Dear Akshay,

Most surge strips contain metal oxyde varistors (MOV) which
are capping the spikes. When they do that they act like
a non-linear devices and create lots of harmonics. But
I'd like to add a word of caution:

If you try several suge strips from different manufacturers
and they all cause this problem that might indicate that
you have some equipment on the line that generates
spikes. Then it may be good to find out what equipment
that is because it can cause damage to lab instruments
that are connected to the line without a surge protector.

I mostly use a little AM radio to find out what causes
spikes or noise.

All the best, Joerg.

----------- Start Original Message -----------
: Dear Joerg:
: Thanks much for the response. You know what? It turned out the surge I was using had a problem with one of the outlet...:-). As soon as I plugged it in a different outlet...BAM...the 60 Hz died. As they say, more than often...the solutions are simple...Thanks again....Akshay
: : Dear Akshay,
: : You'd have to provide some more info about your setup,
: : such as what frequency range you are using or maybe
: : which equipment.
: : Shielding doesn't have much effect at low frequencies
: : and it is usually better to use coupling transformers
: : that pass the desired frequency range and isolate
: : conductive paths. If harmonics are bothering you up
: : into the frequency range you are using then check for
: : noise "polluters" on the mains supply. I found one in
: : our lab yesterday and it was a plain old light dimmer
: : that had gone bad. It was showing up at several MHz
: : and after removing it that noise completely vanished.
: : All the best, Joerg.
: : : Hello:
: : : I am using piezo films for sensing cracks in a substrate. I am having problems with dealing with the 60 Hz supply. It is appearing along with its harmonics in the entire frequency spectrum that I am looking at. I shielded the cables with Al foil, covered the film with Al foil but I am still picking up the 60 Hz. I would highly appreciate any suggestions. Regards.. Akshay
------------ End Original Message ------------




 
 Reply 
 
Ryan Maladen
Ryan Maladen
06:42 Sep-05-2003
Re: PZT film sensing
Dear Akshay/Joerg:
At IIT Delhi we are working on an PZT film sensor which is to be implanted to measure the intracranial pressure of an individual
The problem is we require to measure at a frequency of 1-10 Hz ....with PZT ceramics the same is not possible so cud u give me information regarding the usage of PZT film for my application
Thanks a ton
Looking forward to ur reply
----------- Start Original Message -----------
: Dear Joerg:
: Thanks much for the response. You know what? It turned out the surge I was using had a problem with one of the outlet...:-). As soon as I plugged it in a different outlet...BAM...the 60 Hz died. As they say, more than often...the solutions are simple...Thanks again....Akshay
: : Dear Akshay,
: : You'd have to provide some more info about your setup,
: : such as what frequency range you are using or maybe
: : which equipment.
: : Shielding doesn't have much effect at low frequencies
: : and it is usually better to use coupling transformers
: : that pass the desired frequency range and isolate
: : conductive paths. If harmonics are bothering you up
: : into the frequency range you are using then check for
: : noise "polluters" on the mains supply. I found one in
: : our lab yesterday and it was a plain old light dimmer
: : that had gone bad. It was showing up at several MHz
: : and after removing it that noise completely vanished.
: : All the best, Joerg.
: : : Hello:
: : : I am using piezo films for sensing cracks in a substrate. I am having problems with dealing with the 60 Hz supply. It is appearing along with its harmonics in the entire frequency spectrum that I am looking at. I shielded the cables with Al foil, covered the film with Al foil but I am still picking up the 60 Hz. I would highly appreciate any suggestions. Regards.. Akshay
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

UCI Hardness Tester NOVOTEST T-U2

UCI hardness tester NOVOTEST T-U2 is is used for non-destructive hardness testing of: metals and
...
alloys by scales of hardness: Rockwell (HRC), Brinell (HB), Vickers (HV); non-ferrous metals, alloys of iron etc., and using five additional scales for calibration; with tensile strength (Rm) scale determines the tensile strength of carbon steel pearlitic products by automatic recalculation from Brinell (HB) hardness scale.
>

MIZ®-21C: Truly Affordable Eddy Current Handheld with Surface Array Capability

Introducing MIZ-21C, the first truly affordable handheld eddy current instrument with surface array
...
capabilities. MIZ-21C can deliver fast, accurate inspections in demanding NDT applications including aerospace, oil and gas, manufacturing, and power generation. The surface array solution can reduce inspection time by up to 95% compared to traditional handheld pencil probes. The ergonomic design, long battery life, and intuitive touchscreen mean you can inspect more areas faster than ever without fatigue.
>

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

A1525 SOLO

A1525 Solo – the most compact and affordable TMF unit with two phased array transducers and 3D v
...
isualization and analysis software in standard delivery set. A compact, ergonomic and easy to handheld Phased Array unit based on Total Focusing Method for easy-going imaging of inspection objects with two-dimensional and three-dimensional visualization and evaluation of inspection results.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window