Darlene Johnston
Darlene Johnston
00:40 Apr-30-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: Thanks
------------ End Original Message ------------

1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.



    
 
 
BILL KIESEL
NDT Inspector, level III L.P.I., X-RAY & U.T.
Smiths Tubular Systems-Laconia, USA, Joined Aug 2002, 4

BILL KIESEL

NDT Inspector, level III L.P.I., X-RAY & U.T.
Smiths Tubular Systems-Laconia,
USA,
Joined Aug 2002
4
02:43 Apr-30-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : Thanks
: 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
------------ End Original Message ------------

Most common method for performing TAM panel calibrations require recording the results obtained when the panel is new using new (unused) test materials. Once the size of each indication is recorded the panel can be put into a recall system and the same results must be obtained each time the panel is recalled for calibration. It is important to use the same materials and the same processing parameters. A decrease in indication size will indicate the panel needs cleaning and a increase in size will indicate panel degradation.


    
 
 
Dave Davies
Dave Davies
06:22 May-03-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : Thanks
: 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
------------ End Original Message ------------

What do you mean by "calibration"? The TAM panel, to the best of my knowledge, is a comparative item used to check deterioration of chemicals or other problems in the system on an ongoing basis.

By photographing the new panel with a known good process / consumables a baseline is established against which you can see any deterioration. (You should have two photographs, keeping one for on line display/use and keeping one in a light proof storage as a backup).

Whether deterioration comes from problems with the panel or process is what you have to determine. You should be keeping unused samples of consumables so if you suspect TAM panel problems process it with the saved materials and check against your initial photo.

"Calibration" is thus done by yourself on your system and would not be valid for any other system. Keeping the TAM panel immersed in a solvent between uses and regular vapour / ultrasonic cleaning should slow down the build up of penetrant residue in the cracks on the panel.

You need to write your own procedure for this, suppliers when they "calibrate" panels process with unused materials and check against file photo's, there should be no reason why you can't do the same (provided your regulators and customers agree of course).

Dave Davies


    
 
 
Dave
Dave
02:09 May-04-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : : Thanks
: : 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
: Most common method for performing TAM panel calibrations require recording the results obtained when the panel is new using new (unused) test materials. Once the size of each indication is recorded the panel can be put into a recall system and the same results must be obtained each time the panel is recalled for calibration. It is important to use the same materials and the same processing parameters. A decrease in indication size will indicate the panel needs cleaning and a increase in size will indicate panel degradation.
------------ End Original Message ------------


What do you mean by "calibration"? The TAM panel. to the best of my knowledge,is a comparative item used to check deterioration of chemicals or other problems in the system on an ongoing basis.

By photographing the new panel with a known good process / consumables a baseline is established against which you can see any deterioration. (You should have two photographs, keeping one for on line display/use and keeping one in a light proof storage as a backup).

Whether this comes from problems with the panel or process is what you have to determine. You should be keeping unused samples of consumables so if you suspect TAM panel problems process it with the saved materials and check against your initial photo.

"Calibration" is thus done by yourself on your system and would not be valid for any other system. Keeping the TAM panel immersed in a solvent between uses and regularvapour / ultrasonic cleaning should slow down the build up of penetrant residue in the cracks on the panel.

You need to write your own procedure for this, suppliers when they "calibrate" panels just process with unused materials and check against file photo's, there should be no reason why you can't do the same (provided your regulators and customers agree of course).





    
 
 

Product Spotlight

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

Trade-In Offer

We are offering our existing clients trade-in against Wavemaker® G3 systems, Standard (Blue) inflat
...
able rings and modules for excellent deals on the latest Wavemaker® G4mini system.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window