Re: Calibration of TAM Panel
----------- Start Original Message -----------
: Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: Thanks
------------ End Original Message ------------

1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.



 
 Reply 
 
BILL KIESEL
NDT Inspector, level III L.P.I., X-RAY & U.T.
Smiths Tubular Systems-Laconia, USA, Joined Aug 2002, 4

BILL KIESEL

NDT Inspector, level III L.P.I., X-RAY & U.T.
Smiths Tubular Systems-Laconia,
USA,
Joined Aug 2002
4
02:43 Apr-30-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : Thanks
: 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
------------ End Original Message ------------

Most common method for performing TAM panel calibrations require recording the results obtained when the panel is new using new (unused) test materials. Once the size of each indication is recorded the panel can be put into a recall system and the same results must be obtained each time the panel is recalled for calibration. It is important to use the same materials and the same processing parameters. A decrease in indication size will indicate the panel needs cleaning and a increase in size will indicate panel degradation.


 
 Reply 
 
Dave Davies
Dave Davies
06:22 May-03-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : Thanks
: 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
------------ End Original Message ------------

What do you mean by "calibration"? The TAM panel, to the best of my knowledge, is a comparative item used to check deterioration of chemicals or other problems in the system on an ongoing basis.

By photographing the new panel with a known good process / consumables a baseline is established against which you can see any deterioration. (You should have two photographs, keeping one for on line display/use and keeping one in a light proof storage as a backup).

Whether deterioration comes from problems with the panel or process is what you have to determine. You should be keeping unused samples of consumables so if you suspect TAM panel problems process it with the saved materials and check against your initial photo.

"Calibration" is thus done by yourself on your system and would not be valid for any other system. Keeping the TAM panel immersed in a solvent between uses and regular vapour / ultrasonic cleaning should slow down the build up of penetrant residue in the cracks on the panel.

You need to write your own procedure for this, suppliers when they "calibrate" panels process with unused materials and check against file photo's, there should be no reason why you can't do the same (provided your regulators and customers agree of course).

Dave Davies


 
 Reply 
 
Dave
Dave
02:09 May-04-2004
Re: Calibration of TAM Panel
----------- Start Original Message -----------
: : : Hi, l am from Singapore Airline NDT section. Want to ask how do we calibrate the TAM Panel for FPI ? I know one way is to send back to manufactuer. Other than that any other method we could do on our own?
: : : Thanks
: : 1st problem - does your company allow for in-house calibrations or do you follow the NDT standards - if so, you will have to follow their guidelines. Just to let you know, if you clean your panel (using acetone in an ultrasonic cleaner for approx 2 hrs) you get a VERY clean panel.
: Most common method for performing TAM panel calibrations require recording the results obtained when the panel is new using new (unused) test materials. Once the size of each indication is recorded the panel can be put into a recall system and the same results must be obtained each time the panel is recalled for calibration. It is important to use the same materials and the same processing parameters. A decrease in indication size will indicate the panel needs cleaning and a increase in size will indicate panel degradation.
------------ End Original Message ------------


What do you mean by "calibration"? The TAM panel. to the best of my knowledge,is a comparative item used to check deterioration of chemicals or other problems in the system on an ongoing basis.

By photographing the new panel with a known good process / consumables a baseline is established against which you can see any deterioration. (You should have two photographs, keeping one for on line display/use and keeping one in a light proof storage as a backup).

Whether this comes from problems with the panel or process is what you have to determine. You should be keeping unused samples of consumables so if you suspect TAM panel problems process it with the saved materials and check against your initial photo.

"Calibration" is thus done by yourself on your system and would not be valid for any other system. Keeping the TAM panel immersed in a solvent between uses and regularvapour / ultrasonic cleaning should slow down the build up of penetrant residue in the cracks on the panel.

You need to write your own procedure for this, suppliers when they "calibrate" panels just process with unused materials and check against file photo's, there should be no reason why you can't do the same (provided your regulators and customers agree of course).





 
 Reply 
 

Product Spotlight

Magnetic X-Ray Pipeline Crawler 6''-60''

Suitable to work in 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometer
...
s and provide the efficient inspection of the pipeline.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Robotic laser shearography enables 100% inspection of complex, flight-critical composite structures

An article in “Composites World Magazine” showcases Non Destructive Testing of aero-structures
...
with Laser Shearography. Over the years Dantec Dynamics has supplied many solutions for the aerospace industry. Referring to specific customer projects several of these cases are examined to outline the advantages of using Laser Shearography for automated defect detection.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window