Limits of accurcacy with TOFD

Could anyone let me know the limits on accuracy to be considered when sizing a fatigue crack using TOFD.Can a 0.1" deep crack on a 0.5" thick steel be sized with reasonable accuracy.



 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1304

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1304
00:54 Sep-08-2004
Re: Limits of accurcacy with TOFD
Sri:
In answering your question about limitation of TOFD to "size" a fatigue crack, my reply will be like most NDT technicians' responses...a bit vague. Or perhaps more accurately, cautionary.

The condition of a 0.1" (2.5mm) crack in a 0.5" (12.5mm) wall is not too difficult to assess IF the crack is on the opposite wall from the probes, the plate tested is flat and the crack is surface connected. Typically such a crack "tip" can be located within a half wavelength of the frequency used. For 10MHz compression mode in steel the wavelength is about 0.6mm so an approximate "tolerance" of 0.3mm might be the sort of accuracy you might be able to achieve under good conditions.

But if the flaw is on the near surface this will be in the lateral wave ring time and not reasonably sizeable! Also, if the flaw is not surface connected on the opposite wall the ability to resolve the lower tip will limit the accuracy as the ring time from the lower tip could be masked by the backwall ring.

To some extent the resolution will also be limited by the angle used. Better time resolution is had for smaller angles but these may not be the best option for thin wall.

If your flaw is a simple surface connecting crack won the opposite wall and you can use a reasonably high frequency probe with a short ringtime (broad bandwidth) and about a 60 degree nominal refracted angle the extent to which the flaw has propagated from the surface should be reasonably estimated to less than 0.5mm (0.02").

Resolution and error limitation guides are well explained in EN 583-6.

----------- Start Original Message -----------
: Could anyone let me know the limits on accuracy to be considered when sizing a fatigue crack using TOFD.Can a 0.1" deep crack on a 0.5" thick steel be sized with reasonable accuracy.
------------ End Original Message ------------




 
 Reply 
 
Srikanth Ramachandran
Srikanth Ramachandran
09:00 Sep-08-2004
Thanks for responding: Limits of accurcacy with TOFD
Thank you very much.Its of great help in computing possible errors.

Thanks again
Sri


----------- Start Original Message -----------
: Sri:
: In answering your question about limitation of TOFD to "size" a fatigue crack, my reply will be like most NDT technicians' responses...a bit vague. Or perhaps more accurately, cautionary.
: The condition of a 0.1" (2.5mm) crack in a 0.5" (12.5mm) wall is not too difficult to assess IF the crack is on the opposite wall from the probes, the plate tested is flat and the crack is surface connected. Typically such a crack "tip" can be located within a half wavelength of the frequency used. For 10MHz compression mode in steel the wavelength is about 0.6mm so an approximate "tolerance" of 0.3mm might be the sort of accuracy you might be able to achieve under good conditions.
: But if the flaw is on the near surface this will be in the lateral wave ring time and not reasonably sizeable! Also, if the flaw is not surface connected on the opposite wall the ability to resolve the lower tip will limit the accuracy as the ring time from the lower tip could be masked by the backwall ring.
: To some extent the resolution will also be limited by the angle used. Better time resolution is had for smaller angles but these may not be the best option for thin wall.
: If your flaw is a simple surface connecting crack won the opposite wall and you can use a reasonably high frequency probe with a short ringtime (broad bandwidth) and about a 60 degree nominal refracted angle the extent to which the flaw has propagated from the surface should be reasonably estimated to less than 0.5mm (0.02").
: Resolution and error limitation guides are well explained in EN 583-6.
: : Could anyone let me know the limits on accuracy to be considered when sizing a fatigue crack using TOFD.Can a 0.1" deep crack on a 0.5" thick steel be sized with reasonable accuracy.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)

Download our white paper on Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)
...
Using 3D VSHARP, a fast GPU-Based Linear Boltzmann Transport Equation Solver here: http://bit.ly/2Y0t78P
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window