05:18 Sep-25-2004
Pizo-elecric crystal frequency variation

Hi, Friendes

During the assembling of the ultrasonic probe, nominal frequency of the Pizo-electric crystal get change is it true? What are the factores? How we can minimise that?

Frequency is depend on the thickness of the crystal and less thicness means high frequency.

As per the ASTM 10% is acceptable is it right?


    
 
 
Paul A. Meyer
R & D,
GE Inspection Technologies, USA, Joined Nov 1998, 47

Paul A. Meyer

R & D,
GE Inspection Technologies,
USA,
Joined Nov 1998
47
02:54 Sep-27-2004
Re: Pizo-elecric crystal frequency variation
Hello,
Piezoelectric elements are initially characterized while connected to a network analyzer and having no mechanical loads applied. As they are built into transducers, damping material, matching layers, and electrical tuning components are added to produce the desired performance. Therefore, to optimize the perfromance of the completed probe, the initial piezoelctric element parameters become design variables. For example, to build a highly damped 5 MHz probe, a 6 MHz piezoelectric element may be required to account for mechanical and electrical loads. All components in the probe design need to be considered to optimize probe performance. There are transducer models that can be used for this purpose. Many are based on the KLM transmission line concept published many years ago. These are very helpful but are 1-dimensional, lateral effects are not considered. FEA analysis is more detailed but more time comsuming and computationally intensive.
Paul

----------- Start Original Message -----------
: Hi, Friendes
: During the assembling of the ultrasonic probe, nominal frequency of the Pizo-electric crystal get change is it true? What are the factores? How we can minimise that?
: Frequency is depend on the thickness of the crystal and less thicness means high frequency.
: As per the ASTM 10% is acceptable is it right?
------------ End Original Message ------------




    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1219

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1219
01:58 Sep-27-2004
Re: Pizo-elecric crystal frequency variation
Paul:
Your mention of how "loading" reduces the ideal centre frequency determined by the frequency constant is well worth noting. I have also noted that it is very important to define the method of frequency assessment.
Should you use pulse-echo or transmit-receive? Should the assessment be made in water or in the material that you are designing the probe to operate in?
Be careful if you are mounting the element on a refracting wedge. The wedge material will restrict frequency content.
At what distance should you measure the frequency?
If using a digital scope for analysis what is the digitising frequency that should be used?
Finally, there are several "frequencies" that may be used to describe the probe; design, nominal, centre, peak, or upper (or lower) 6dB edge. (I once bought a 50MHz PVDF probe that I could not get to pulse with a peak frequency of better than about 28MHz. The upper and lower 6dB were about 50 and 20MHz so some components were at the "nominal" but I did not understand how it was marketed at 50MHz.

There are also pulse characteristics to consider. A spike pulser will not produce the same frequency content as a square-wave or chopped sine-wave pulser.
Regards
Ed

----------- Start Original Message -----------
: Hello,
: Piezoelectric elements are initially characterized while connected to a network analyzer and having no mechanical loads applied. As they are built into transducers, damping material, matching layers, and electrical tuning components are added to produce the desired performance. Therefore, to optimize the perfromance of the completed probe, the initial piezoelctric element parameters become design variables. For example, to build a highly damped 5 MHz probe, a 6 MHz piezoelectric element may be required to account for mechanical and electrical loads. All components in the probe design need to be considered to optimize probe performance. There are transducer models that can be used for this purpose. Many are based on the KLM transmission line concept published many years ago. These are very helpful but are 1-dimensional, lateral effects are not considered. FEA analysis is more detailed but more time comsuming and computationally intensive.
: Paul
: : Hi, Friendes
: : During the assembling of the ultrasonic probe, nominal frequency of the Pizo-electric crystal get change is it true? What are the factores? How we can minimise that?
: : Frequency is depend on the thickness of the crystal and less thicness means high frequency.
: : As per the ASTM 10% is acceptable is it right?
------------ End Original Message ------------




    
 
 
Amry Amin
R & D
Malaysian Inst. for Nuclear Technology Research, Malaysia, Joined Aug 2002, 4

Amry Amin

R & D
Malaysian Inst. for Nuclear Technology Research,
Malaysia,
Joined Aug 2002
4
00:39 Sep-29-2004
Re: Pizo-elecric crystal frequency variation
Hello,

Does backing material, front face material of the disc and load of the medium have effect on the frequency of a complete probe? I have made a model of a probe considering the layers using Mason circuit and I found that the resonant frequency varies when the parameters of the layers are altered.

----------- Start Original Message -----------
: Hello,
: Piezoelectric elements are initially characterized while connected to a network analyzer and having no mechanical loads applied. As they are built into transducers, damping material, matching layers, and electrical tuning components are added to produce the desired performance. Therefore, to optimize the perfromance of the completed probe, the initial piezoelctric element parameters become design variables. For example, to build a highly damped 5 MHz probe, a 6 MHz piezoelectric element may be required to account for mechanical and electrical loads. All components in the probe design need to be considered to optimize probe performance. There are transducer models that can be used for this purpose. Many are based on the KLM transmission line concept published many years ago. These are very helpful but are 1-dimensional, lateral effects are not considered. FEA analysis is more detailed but more time comsuming and computationally intensive.
: Paul
: : Hi, Friendes
: : During the assembling of the ultrasonic probe, nominal frequency of the Pizo-electric crystal get change is it true? What are the factores? How we can minimise that?
: : Frequency is depend on the thickness of the crystal and less thicness means high frequency.
: : As per the ASTM 10% is acceptable is it right?
------------ End Original Message ------------




    
 
 
Paul A. Meyer
R & D,
GE Inspection Technologies, USA, Joined Nov 1998, 47

Paul A. Meyer

R & D,
GE Inspection Technologies,
USA,
Joined Nov 1998
47
03:43 Oct-01-2004
Re: Pizo-elecric crystal frequency variation
Hello Amry,
The reflection and transmission factors for a thin layer depend on the wavelength-to-thickness ratio. Most ultrasound transducers emit pulses with fractional bandwidths greater than 40%. The apparent mechanical load "seen" by the ceramic through the face will depend on the wavelength (frequency). If the thickness of the layer is changed, the variation of load with wavelength also changes, which changes the response of the probe. You are approaching the problem correctly by modeling the design for your particular need. You will find that multiple face layers can improve probe performance in many situations.
Paul
----------- Start Original Message -----------
: Hello,
: Does backing material, front face material of the disc and load of the medium have effect on the frequency of a complete probe? I have made a model of a probe considering the layers using Mason circuit and I found that the resonant frequency varies when the parameters of the layers are altered.
: : Hello,
: : Piezoelectric elements are initially characterized while connected to a network analyzer and having no mechanical loads applied. As they are built into transducers, damping material, matching layers, and electrical tuning components are added to produce the desired performance. Therefore, to optimize the perfromance of the completed probe, the initial piezoelctric element parameters become design variables. For example, to build a highly damped 5 MHz probe, a 6 MHz piezoelectric element may be required to account for mechanical and electrical loads. All components in the probe design need to be considered to optimize probe performance. There are transducer models that can be used for this purpose. Many are based on the KLM transmission line concept published many years ago. These are very helpful but are 1-dimensional, lateral effects are not considered. FEA analysis is more detailed but more time comsuming and computationally intensive.
: : Paul
: : : Hi, Friendes
: : : During the assembling of the ultrasonic probe, nominal frequency of the Pizo-electric crystal get change is it true? What are the factores? How we can minimise that?
: : : Frequency is depend on the thickness of the crystal and less thicness means high frequency.
: : : As per the ASTM 10% is acceptable is it right?
------------ End Original Message ------------




    
 
 
Dan Walden
Dan Walden
07:07 Feb-21-2005
Re: Pizo-elecric crystal frequency variation
I have a fountain that has water and gas flames. I would like to know if there is some way to design some way to create an electrical lighter, so flame is kept burning.





    
 
 
nitin joshi
nitin joshi
06:12 Mar-02-2005
miro piezoelectric crystal
plese send me complete tutorial about micro piezo electric crystal


    
 
 
Bhavesh
Bhavesh
04:01 Dec-31-2005
Re: Pizo-elecric crystal frequency variation
----------- Start Original Message -----------
hello,
i have no idea for ultrasonic welding in used transuder details for pizo. i have make for 20khz frequency what about parametaer of pizo design and what about size and what about attachment to pizo for make trasuder for 2o khz.


    
 
 

Product Spotlight

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

TVC awarded UKAS accreditation

TVC are delighted to finally announce we have been awarded UKAS accreditation for our calibration
...
laboratory. Laboratory accreditation to ISO/IEC 17025:2005 enables us to conduct the Electrical Verification of Ultrasonic Flaw Detection Equipment to BS EN 12668-1:2010. It has taken many months of hard work and we want to thank our staff for all their efforts during this massive undertaking.
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

MUSE Mobile Ultrasonic Equipment