Flatness of ut probe face

Hi ,


In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.


    
 
 Reply 
 
Phil Herman Jr.
Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards, USA, Joined Oct 1999, 79

Phil Herman Jr.

Sales, - Manufacture of NDT Reference Standards/Test Blocks
PH Tool Reference Standards,
USA,
Joined Oct 1999
79
09:35 Dec-07-2004
Re: Flatness of ut probe face
Amry,

It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.


----------- Start Original Message -----------
: Hi ,
:
: In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




    
 
 Reply 
 
DREW
DREW
01:08 Sep-02-2005
Re: Flatness of ut probe face
----------- Start Original Message -----------
: Amry,
: It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.
:
: : Hi ,
: :
: : In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




    
 
 Reply 
 
DREW
DREW
01:08 Sep-02-2005
Re: Flatness of ut probe face
----------- Start Original Message -----------
: Amry,
: It will be difficult using a ruler or feeler gage to determine if the flatness over the probe face is less than 0.05mm (0.002"). To perform this check accurately, you need a dial indicator (aka test indicator). This is a common inspection tool used in the toolmaking trades, that measures flatness and other dimensional features by the amount of movement of the indicating arm. The movement, or variation is displayed on a clock-like dial which reads in MMs or thousandths of an inch. Set up the probe in a small vise or vee-block with the face up, and place the indicator in a surface gage to hold it. Move the surface gage base over a granite surface plate or other flat surface so that the indicator arm travels over the probe face. Set a 'zero' on the indicator by rotating the ring to 0. Observe the minumum and maximum readings and ensure the total indicator reading (TIR) does not exceed the allowable flatness.
:
: : Hi ,
: :
: : In EN12668-2, it is mentioned that to measure flatness of contact probes use a ruler or feeler gauges. The gap over the whole probe face shall not be larger than 0.05mm. Can anyone please advice on the method to be used. Thanks.
------------ End Original Message ------------




    
 
 Reply 
 

Product Spotlight

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

XRHRobotStar

In high volume industries like automotive the requirement for a hundred percent X-ray inspection c
...
reates a bottleneck in the production. The XRHRobotStar is a fully Automated Defect Recognition (ADR) capable robot-system that allows an ultra-fast in-line inspection.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window