Up to now we used our own testing methods. Because thus the values are not good comparable I`m searching
for standarized testing methods for our ultrasonic transducers.
Who know standards (e. g. ASTM, which number?), which describes testing methods for the specification of
ultrasonic transducers? I`m searching for standardized methods for the determination of:
- Sensitivity (e. g. P/E-measurement, which reflector distance?, what driving signal (burst with how many cycles or pulse...)
- Pulse width (How must the driving pulse be specified for a standard measurement?)
- Bandwidth (Is in the standard method a pulse measurement with Fourier transformation used, or is a sweeped CW signal used?)



 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
01:15 May-11-1999
Re: Standards for Testing Ultrasonic Transducers

: Up to now we used our own testing methods. Because thus the values are not good comparable I`m searching
: for standarized testing methods for our ultrasonic transducers.
: Who know standards (e. g. ASTM, which number?), which describes testing methods for the specification of
: ultrasonic transducers? I`m searching for standardized methods for the determination of:
: - Sensitivity (e. g. P/E-measurement, which reflector distance?, what driving signal (burst with how many cycles or pulse...)
: - Pulse width (How must the driving pulse be specified for a standard measurement?)
: - Bandwidth (Is in the standard method a pulse measurement with Fourier transformation used, or is a sweeped CW signal used?)

You should obtain a copy of ASTM E 1065-96, "Standard Guide for Evaluating Characteristics of Ultrasonic Search Units". It is fairly detailed, and seems to be generally recognized as a good standard for NDT transducer qualification.

--Tom Nelligan
Senior Applications Engineer


 
 Reply 
 
Rolf D.
Director,
NDT.net, Germany, Joined Nov 1998, 618

Rolf D.

Director,
NDT.net,
Germany,
Joined Nov 1998
618
01:47 May-11-1999
Re: Standards for Testing Ultrasonic Transducers
al Committee
CEN/TC 138 "Non-destructive testing" of which the secretariat is held by AFNOR.

This document is currently submitted to the formal vote.

This standard consists of the following parts:
EN 12668-1, Non-destructive testing and verification of ultrasonic
examination equipment - Part 1: Instruments
EN 12668-2, Non-destructive testing - Characterization and
verification of ultrasonic examination equipment - Part 2: Probes
EN 12668-3, Non-destructive testing - Characterization and
verification of ultrasonic examination equipment - Part 3: Combined
equipment


Rolf Diederichs



 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1304

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1304
06:29 May-16-1999
Re: Standards for Testing Ultrasonic Transducers
In addition to the references by Tom & Rolf DIN 25 450
is a good and comprehensive Code.
Also see:
AFNOR A09-323
JIS Z 2352
ESI-98-9
BS 4331

Ed


: Up to now we used our own testing methods. Because thus the values are not good comparable I`m searching
: for standarized testing methods for our ultrasonic transducers.
: Who know standards (e. g. ASTM, which number?), which describes testing methods for the specification of
: ultrasonic transducers? I`m searching for standardized methods for the determination of:
: - Sensitivity (e. g. P/E-measurement, which reflector distance?, what driving signal (burst with how many cycles or pulse...)
: - Pulse width (How must the driving pulse be specified for a standard measurement?)
: - Bandwidth (Is in the standard method a pulse measurement with Fourier transformation used, or is a sweeped CW signal used?)




 
 Reply 
 

Product Spotlight

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

On Demand Webinar: Using Eddy Current Array for Large Areas in Place of Magnetic Particle and Liquid Penetrant Testing

New on-demand webinar on the use of eddy current array (ECA) technology in lieu of liquid penetrant
...
testing (PT) and magnetic particle testing (MT) for large area screenings. The webinar includes an introduction to the technologies with advantages and disadvantages of each. Side by side examples are discussed including aircraft stringers, engine blades, forgings, weld plates & more. View it today.
>

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window