Mathias
Mathias
01:18 May-25-2005
ASTM E 1742-00

Can anyone explain why the curve in Figure 2 (Maximum Allowable Film Density) isn’t continous? There is a jump between density 2.5 and 3.0 – why?



    
 
 Reply 
 
DJ Kallhof
DJ Kallhof
04:25 May-31-2005
Re: ASTM E 1742-00
I believe they are attempting to illustrate the difference between the larger "low range " viewer and the smaller "high intensity" viewer. This is just a way to document and justify the maximum readable film density posted on each viewer. What I find interesting is the fact we can no longer measure the light intensity in foot candles. According to ASTM E 1742-05 we must now measure luminance in candelas/m squared. Again, the cost of NDT goes up. Yee Ha
----------- Start Original Message -----------
: Can anyone explain why the curve in Figure 2 (Maximum Allowable Film Density) isn’t continous? There is a jump between density 2.5 and 3.0 – why?
:
------------ End Original Message ------------




    
 
 Reply 
 
Mathias
Mathias
05:17 Jun-01-2005
Re: ASTM E 1742-00
Thank you for the answer Mr Kallhof. But what still makes me confused is the fact that if you now measure 10000 lux at the viewing port, can you post a 3.0 sticker on the viewer or is the maximum readable density just 2.5?
----------- Start Original Message -----------
: I believe they are attempting to illustrate the difference between the larger "low range " viewer and the smaller "high intensity" viewer. This is just a way to document and justify the maximum readable film density posted on each viewer. What I find interesting is the fact we can no longer measure the light intensity in foot candles. According to ASTM E 1742-05 we must now measure luminance in candelas/m squared. Again, the cost of NDT goes up. Yee Ha
: : Can anyone explain why the curve in Figure 2 (Maximum Allowable Film Density) isn’t continous? There is a jump between density 2.5 and 3.0 – why?
: :
------------ End Original Message ------------




    
 
 Reply 
 
DJ Kallhof
DJ Kallhof
09:27 Jun-03-2005
Re: ASTM E 1742-00
Unfortunately I'm in the dark over this revision as much as yourself. I am waiting for a response from ASTM subcommittee E07.01 to my inquiry regarding the posting of "Maximum Allowable Film Density" and how to come up with a value.


----------- Start Original Message -----------
: Thank you for the answer Mr Kallhof. But what still makes me confused is the fact that if you now measure 10000 lux at the viewing port, can you post a 3.0 sticker on the viewer or is the maximum readable density just 2.5?
: : I believe they are attempting to illustrate the difference between the larger "low range " viewer and the smaller "high intensity" viewer. This is just a way to document and justify the maximum readable film density posted on each viewer. What I find interesting is the fact we can no longer measure the light intensity in foot candles. According to ASTM E 1742-05 we must now measure luminance in candelas/m squared. Again, the cost of NDT goes up. Yee Ha
: : : Can anyone explain why the curve in Figure 2 (Maximum Allowable Film Density) isn’t continous? There is a jump between density 2.5 and 3.0 – why?
: : :
------------ End Original Message ------------




    
 
 Reply 
 

Product Spotlight

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

Robotic laser shearography enables 100% inspection of complex, flight-critical composite structures

An article in “Composites World Magazine” showcases Non Destructive Testing of aero-structures
...
with Laser Shearography. Over the years Dantec Dynamics has supplied many solutions for the aerospace industry. Referring to specific customer projects several of these cases are examined to outline the advantages of using Laser Shearography for automated defect detection.
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window