02:07 Jun-10-2005
ASTM E1742-05

With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?


    
 
 
Mathias
Mathias
08:43 Jun-13-2005
Re: ASTM E1742-05
I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
----------- Start Original Message -----------
: With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




    
 
 
DJ Kallhof
DJ Kallhof
07:40 Jun-22-2005
Re: ASTM E1742-05
Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.

----------- Start Original Message -----------
: I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




    
 
 
Mathias
Mathias
00:55 Jun-27-2005
Re: ASTM E1742-05
Yes, seems correct. But I hope you don´t mean currently availabe on the market?!

----------- Start Original Message -----------
: Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.
: : I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




    
 
 

Product Spotlight

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

X-ray CT aids research into defect formation in AM parts

X-ray CT is used to research how additive manufacturing process parameters influence defect format
...
ion in AM parts.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window