ASTM E1742-05

With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?


 
 Reply 
 
Mathias
Mathias
08:43 Jun-13-2005
Re: ASTM E1742-05
I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
----------- Start Original Message -----------
: With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 
DJ Kallhof
DJ Kallhof
07:40 Jun-22-2005
Re: ASTM E1742-05
Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.

----------- Start Original Message -----------
: I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 
Mathias
Mathias
00:55 Jun-27-2005
Re: ASTM E1742-05
Yes, seems correct. But I hope you don´t mean currently availabe on the market?!

----------- Start Original Message -----------
: Light meters currently available register up to 19,990 lux. That level will only qualify the film viewer to interprate 3.30 density film, according to ASTM E 1742-05 Fig. 2. If I am incorrect thinking this please respond.
: : I believe you have to purchase a new lightmeter. The new requirement applies to the excisting viewers whenever the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used. This is my interpretation.
: : : With reference to Para 6.27.4.4, it states that the "maximum readable film density values shall be established when the viewer is procured and shall be re-establish when the viewer is repaired, moved, altered, or a different bulb, bulb designation or manufacturer is used".
: : : Most of us are using our existing film viewers, in which the maximum film density values are checked using a light meter to measure transmitted light, in foot candles, through a 2.0D film, in accordance with ASTM E1742-00. If we do not intend to procure a new viewer, do we still have to purchase a new lightmeter to measure the light intensity in candelas/m squared, according to ASTM E1742-05? Am I right to say that it only applies to newly procured viewers?
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

GUL QSR1® Scanning

How do you measure pipe wall thickness without direct access to the area? QSR® Scanning - Guide Wav
...
e Quantitative Short Range Scanning.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window