Near Field for TR Probes

How to calculate Near Field lenght for TR Probes?


    
 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1234

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1234
05:23 Aug-29-2005
Re: Near Field for TR Probes
Near field is a result of interference effects from the pulsed tranducer surface. Therefore the near field would apply to the transmitter. In the receive mode a piezo element is passive and sound pressure must impinge on it to generate a voltage indicated as the "received signal" so the interference effects associated with the near field are not applicable.

Assuming a single element for each of the transmit and receive sides of the probe, the near zone calculation should apply using the standard calculations for the transmitter element only.

Ed

----------- Start Original Message -----------
: How to calculate Near Field lenght for TR Probes?
------------ End Original Message ------------




    
 
 
Devendra
Devendra
18:50 Dec-25-2014
Re: Near Field for TR Probes
In Reply to Ed Ginzel at 05:23 Aug-29-2005 .

it means that near field is not applicable for TR and Normal proble please ensure me sir.

    
 
 
Shank Vagal
India, Joined Jun 2014, 73

Shank Vagal

India,
Joined Jun 2014
73
23:57 Dec-25-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

In case of a TR probe you have no near field nor any dead zone. you can start reading from zero of the scale.

    
 
 
mark
mark
00:26 Dec-26-2014
Re: Near Field for TR Probes
In Reply to Devendra at 18:50 Dec-25-2014 .

Ed is right and I want to add some thing more,
TR probes has focal spot and and it is adjustable with roof angle of transmitter and receiver crystal by manufacturer design to get maximum response of reflector at desire focal point.focal spot is also affecte by near field of each crystal and as you know there is maximum response at the end of nearfield of single crystal so focal spot should be set at end of near field of transmitter crystal by manufacturers.you can measures focal spot for TR probes and near field for single probes by making DAC curve with special reference block.

    
 
 
John Brunk
Engineering, NDT Level III
Self employed, part-time, USA, Joined Oct 1999, 158

John Brunk

Engineering, NDT Level III
Self employed, part-time,
USA,
Joined Oct 1999
158
03:43 Dec-29-2014
Re: Near Field for TR Probes
In Reply to Shank Vagal at 23:57 Dec-25-2014 .

Although a TR probe does not exhibit a near field, it may have a dead zone. A small reflector such as a corrosion pit that is too close to the test surface will not necessarily produce a reflection that is well detected by the receiving element. Some TR probes are stated by the manufacturer to have an optimum range, with the minimum and maximum useful limits being somewhat beyond these. If the object is to measure very thin sections a single-element probe with a delay line may be better. But with this type of probe there is less sensitivity to corrosion pits because the beam is spreading beyond the near field, where with a TR robe the beam has effectively a focused zone.

    
 
 
John Norman
Consultant, owner of business
NTS Ultrasonics Pty Ltd, Australia, Joined Oct 2012, 108

John Norman

Consultant, owner of business
NTS Ultrasonics Pty Ltd,
Australia,
Joined Oct 2012
108
09:44 Dec-29-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

Hi Biju.

Mark is correct in that a good way to work it out is to generate a DAC curve for the probe. My concern with calculation based on the transmitter element is that TR probe elements usually have a complex shape, often a "half moon" or something like that. Nearfield calculations using simple text book formula are a bit difficult to apply, even for TR probes with rectangular elements.

Regards
John Norman

    
 
 
Anderson Hwang
Sales, - -
Guangdong Goworld Co., Ltd., China, Joined Sep 2010, 8

Anderson Hwang

Sales, - -
Guangdong Goworld Co., Ltd.,
China,
Joined Sep 2010
8
03:42 Dec-31-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

Dear Sir,

Yes, the peers above my "floor" are all right. And I think both DAC and DGS are OK. DGS might be simple to witness.

    
 
 

Product Spotlight

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

TOPAZ® 64: Fully Integrated Portable 64 Channel PAUT Instrument with FMC & TFM

Introducing TOPAZ64, the industry’s most intelligent fully integrated, portable 64-channel phased
...
array ultrasound (UT) instrument. TOPAZ64 combines code-compliant phased array UT with the industry’s most advanced full matrix capture (FMC) and total focusing method (TFM) capabilities. Featuring the highest acquisition frequency in its class, high resolution FMC and a 12” multi-touchscreen, users can easily visualize even the smallest flaws. TOPAZ64 can generate a bipolar pulse that provides more acoustic energy versus previous models for punching through thick components. The result is a portable tool that delivers increased inspection coverage, more accurate signals, and the ability to handle all UT inspections in one package. TOPAZ64 is ideal for challenging applications in transportation, oil and gas, manufacturing and power generation.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window