Biju Varghese
Biju Varghese
06:54 Aug-19-2005
Near Field for TR Probes

How to calculate Near Field lenght for TR Probes?


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1300

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1300
05:23 Aug-29-2005
Re: Near Field for TR Probes
Near field is a result of interference effects from the pulsed tranducer surface. Therefore the near field would apply to the transmitter. In the receive mode a piezo element is passive and sound pressure must impinge on it to generate a voltage indicated as the "received signal" so the interference effects associated with the near field are not applicable.

Assuming a single element for each of the transmit and receive sides of the probe, the near zone calculation should apply using the standard calculations for the transmitter element only.

Ed

----------- Start Original Message -----------
: How to calculate Near Field lenght for TR Probes?
------------ End Original Message ------------




 
 Reply 
 
Devendra
Devendra
18:50 Dec-25-2014
Re: Near Field for TR Probes
In Reply to Ed Ginzel at 05:23 Aug-29-2005 .

it means that near field is not applicable for TR and Normal proble please ensure me sir.

 
 Reply 
 
Shank Vagal
India, Joined Jun 2014, 73

Shank Vagal

India,
Joined Jun 2014
73
23:57 Dec-25-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

In case of a TR probe you have no near field nor any dead zone. you can start reading from zero of the scale.

 
 Reply 
 
mark
mark
00:26 Dec-26-2014
Re: Near Field for TR Probes
In Reply to Devendra at 18:50 Dec-25-2014 .

Ed is right and I want to add some thing more,
TR probes has focal spot and and it is adjustable with roof angle of transmitter and receiver crystal by manufacturer design to get maximum response of reflector at desire focal point.focal spot is also affecte by near field of each crystal and as you know there is maximum response at the end of nearfield of single crystal so focal spot should be set at end of near field of transmitter crystal by manufacturers.you can measures focal spot for TR probes and near field for single probes by making DAC curve with special reference block.

 
 Reply 
 
John Brunk
Engineering, NDT Level III
Self employed, part-time, USA, Joined Oct 1999, 162

John Brunk

Engineering, NDT Level III
Self employed, part-time,
USA,
Joined Oct 1999
162
03:43 Dec-29-2014
Re: Near Field for TR Probes
In Reply to Shank Vagal at 23:57 Dec-25-2014 .

Although a TR probe does not exhibit a near field, it may have a dead zone. A small reflector such as a corrosion pit that is too close to the test surface will not necessarily produce a reflection that is well detected by the receiving element. Some TR probes are stated by the manufacturer to have an optimum range, with the minimum and maximum useful limits being somewhat beyond these. If the object is to measure very thin sections a single-element probe with a delay line may be better. But with this type of probe there is less sensitivity to corrosion pits because the beam is spreading beyond the near field, where with a TR robe the beam has effectively a focused zone.

 
 Reply 
 
John Norman
Consultant, owner of business
NTS Ultrasonics Pty Ltd, Australia, Joined Oct 2012, 117

John Norman

Consultant, owner of business
NTS Ultrasonics Pty Ltd,
Australia,
Joined Oct 2012
117
09:44 Dec-29-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

Hi Biju.

Mark is correct in that a good way to work it out is to generate a DAC curve for the probe. My concern with calculation based on the transmitter element is that TR probe elements usually have a complex shape, often a "half moon" or something like that. Nearfield calculations using simple text book formula are a bit difficult to apply, even for TR probes with rectangular elements.

Regards
John Norman

 
 Reply 
 
Anderson Hwang
Sales, - -
Guangdong Goworld Co., Ltd., China, Joined Sep 2010, 8

Anderson Hwang

Sales, - -
Guangdong Goworld Co., Ltd.,
China,
Joined Sep 2010
8
03:42 Dec-31-2014
Re: Near Field for TR Probes
In Reply to Biju Varghese at 06:54 Aug-19-2005 (Opening).

Dear Sir,

Yes, the peers above my "floor" are all right. And I think both DAC and DGS are OK. DGS might be simple to witness.

 
 Reply 
 

Product Spotlight

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window