07:58 May-26-2006
Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit


We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.

Thanks & Regards


 
 Reply 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
06:59 May-27-2006
Re: Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit
Jamil

Calibration of the Emcol unit will primarily be to verify the accuacy of the mounted ammeters, so you will need someone who can check their accuracy back to National Standards. For the magnetic flux function which ia a %age, no calibration is feasible, just a function test. Best bet is to find an approved calibration authority in your area.

Hope this helps. Nigel

----------- Start Original Message -----------
:
: We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
: We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.
: Thanks & Regards
------------ End Original Message ------------




 
 Reply 
 
mandar vinze
Engineering, manufacturer
magnafield controls, India, Joined May 2006, 3

mandar vinze

Engineering, manufacturer
magnafield controls,
India,
Joined May 2006
3
02:51 Jun-07-2006
Re: Re-calibration of EMCOL 6K Portable Current Flow Magnetic Particle Inspection Unit
we are manufacturers of M T equipment and would be glad to help you in any matter relating to equipment,accesories ,calliberation service etc.We are in India and are in this for more than 30 yrs----------- Start Original Message -----------
: Jamil
: Calibration of the Emcol unit will primarily be to verify the accuacy of the mounted ammeters, so you will need someone who can check their accuracy back to National Standards. For the magnetic flux function which ia a %age, no calibration is feasible, just a function test. Best bet is to find an approved calibration authority in your area.
: Hope this helps. Nigel
: :
: : We are using the Current Flow Magnetic Particle Inspection Unit. Its re-calibration is overdue and we want to re-calibrate it. Can anyone guide me how to re-calibrate (in-house) or help me in developing the re-calibration procedure.
: : We have some other equipment as digital fiield strength indicator which is calibrated, may be it will help us.
: : Thanks & Regards
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Panther

M2M PANTHER is a phased-array equipment designed for both desktop and industrial applications, offer
...
ing unparalleled performance in a compact unit. It combines the speed required for industrial integrated Phased-Array Ultrasound (PAUT) solutions, with the most complete set of total focusing method (TFM) imaging techniques, making it the ultimate tool for R&D and procedure qualification.
>

Echomac® Small

Available with up to eight channels of electronics to detect and evaluate thickness, flaws and eccen
...
tricity, this UT tester is housed in a convenient, smaller cabinet. This instrument can be used in conjunction with bubbler or immersion tank systems, or with a test bench or in laboratory applications. The Echomac® Small is available in the FD4, FD6 or FD6A versions.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window