A problem with Magnetic Testing

Hi,
On welding of tube, 5ox5mm with two differents materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
Can you give us an explanation, about the origin of this problem?
Best regards, waiting for solution to this problem.


    
 
 Reply 
 
Michael Trinidad
Consultant,
LMATS Pty Ltd , Australia, Joined Jan 2003, 138

Michael Trinidad

Consultant,
LMATS Pty Ltd ,
Australia,
Joined Jan 2003
138
09:41 Jul-09-1999
Re: A problem with Magnetic Testing
Hi,
On welding of tube, 5ox5mm with two different materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
Can you give us an explanation, about the origin of this problem?
Best regards, waiting for solution to this problem.


This can be quite a common problem if there is a significant permeability difference between the two steels. What you are most likely seeing is the flux leakage occurring at the permeability change

Michael Trinidad
ETRS NDT Project Supervisor
KBJV Worsley Expansion Project
Direct Tel: 61-8-9781-6232
Fax: 61-8-9781-6233
Email: etrs@kbjv.com




    
 
 Reply 
 
David Hermanutz
Consultant,
Hbndt.com, China, Joined Jul 2012, 85

David Hermanutz

Consultant,
Hbndt.com,
China,
Joined Jul 2012
85
04:13 Jul-11-1999
Re: A problem with Magnetic Testing
: Hi,
: On welding of tube, 5ox5mm with two different materials (SA213T91 and SA213T22)we have a linear indication (longitudinal),from Magnetic Testing.
: Can you give us an explanation, about the origin of this problem?
: Best regards, waiting for solution to this problem.

:
: This can be quite a common problem if there is a significant permeability difference between the two steels. What you are most likely seeing is the flux leakage occurring at the permeability change

: Michael Trinidad
: ETRS NDT Project Supervisor
: KBJV Worsley Expansion Project
: Direct Tel: 61-8-9781-6232
: Fax: 61-8-9781-6233
: Email: etrs@kbjv.com

It is best to follow up such indications with a dye penetrant procedure to determine the absence of a weld defect.
Or switch to a non magnetic surface test procedure.
In one instance I came across a faulty weld procedure after doing a PMI out of curiosity of such an indication.
148 cut outs and 26 pups later the problem was corrected... (six day overrun on a shutdown due to one typo)

David Hermanutz
Pilbara NDT
Melbourne, Australia





    
 
 Reply 
 

Product Spotlight

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window